Whether its a large field of view for abdominal aortic aneurysm repair, excellent low-dose imaging for interventional pain management or small anatomical detail for cochlear implants, the BV Pulseras innovative feature set helps you get the job done.Excellent low-dose imaging for interventional pain management.
FORUMSView All (147)
-morobo hospital
Thu Jul 09 2026
major components..names and pictures
need buy a component...dont know the name..can identity with a picture
Equipment: bv pulsera c arm philips
-VicMayal
Wed Jun 10 2026
Total voltage of battery set is just +339 Vdc. Battery set was charged during a 24 -hours cicle. Besides that got M326 and M050 errors
Hello, dear members of forum.
I have a Philips Pulsera Mobile C-arm system. Recently, I have been facing up some error codes. It is M326 and M050. After checking it, I found the fan in ESU unit stuck. It was already replaced, but equipment is not working yet.
The whole battery set was recently replaced, but after a 24-hours charging cicle, battery voltage is just +339 Vdc. I know that it should be +360 Vdc. Could someone help me to solve out this issue? I´ll really appreciate any help. Thanks in advance.
Regards!
Equipment: Philips PULSERA mobile C-Arm
-VicMayal
Mon Jan 12 2026
Blurred images when making fluoroscopy
Hello, dear group.
I´m servicing to a C-Arm Pulsera from Philips. When making fluorospopy get blurred images. I´ve already checked voltages on image intensifier power supply. Voltage measured at test point Vg2 is +3.15v, it does not change eventhoug I change format to 17 cm or 23 cm. But according to technical specification on service manual, it should be +0.8v. Those grid voltage values make me think that II power supply is damaged, but I´m not sure about that. Could anybody help me to make a better diagnostic? Your assistance will be unvaluable.
Thanks before hand.
Regards
Equipment: Philips Pulsera C-Arm
  • Pulsed acquisition (30 pulses/sec) for better images at a lower dose
  • Rotating anode power to penetrate larger patients
  • Top quality 3D imaging
  • 9” or 12” image intensifier
  • Excellent maneuverability
X-ray Generator Power Rating7.5 Kw @ 120 kVp
Motorized Tilting/Pivot Rotation360 C-Arm