Quinton - Q-3000
EKG/ECG
The Q 3000 stress console controls the treadmill speed and grade by means of the selected protocol and displays speed and grade information in the numeric display. ECG data is collected from electrode sites on the body. Computes heart rate, ST parameters, and other information. Hardcopy print of ECG and computed information.
- Frequency response: 0.05 to 100 Hz (+1,-3dB) for chart recorder
- Fixed gain control: 5,10,and 20 mm/mV
- Filters: 25 Hz, 4-pole, low pass artifact filter for the screen and a 25 Hz, low pass digital filter for the chart recorder
- Overall signal error: for a ±0.5 to ±5.0 mV signal, error is less than ±10% or 50 µV, whichever is greater
- Input impedance: 10 megohms minimum
- Calculated parameters:heart rate, ST level, and ST slope
- Records three 5 mm wide ECG channels and a 45 mm wide half channel containing alphanumeric data for labelling and calculated parameters
EKG/ECG Type | Stress |
Height | 62.3 in |
Length | 30.5 in |
Weight | 300 lbs |
Image Display | 9 x 7 in |
Width | 21.2 in |
- Paper capacity: 200 feet (61 m) of z-fold paper
- Paper size: 8 ½” x 11″ ( 216 x 280 mm)
- Paper type: thermal paper, z-fold