The issue appears to be within the endoscopy tower unit itself, specifically the processor and light source.

Upon connecting the scope, we observed the following faults:
    1.    Excessive brightness – The camera light is extremely bright and does not auto-adjust. Manual adjustment is required to make the image viewable.
    2.    White balance malfunction – While the system confirms “White balance OK,” the image remains overly bright.
    3.    Iris brightness control – This function does not work.
    4.    Image distortion – Noticeable distortion was observed during twsting of the scopes.

For testing, we used both a 180 series colonoscope and gastroscope, and the issue persisted across both, ruling out the scopes themselves as the source of the fault.

Additionally, we thoroughly checked and cleaned all contact points to ensure there were no connection issues. Despite these efforts, the problem remained unresolved.

Equipment: CV180

RE: Olympus - Exera II CV-180
1 Reply

Hi, has anyone had trouble using the ERBE VIO 100 C with a pencil activated synchronised smoke evacuator?

RE: ERBE - VIO 100 C
Updated  Jul 26 2025
1 Reply

I lost the manual of OPMI Proergo Zeiss. I Would like a copy

Equipment: Microscópio OPMI Proergo Zeiss

RE: ZEISS - OPMI PROergo
Updated  Jul 24 2025
1 Reply

I am looking for the controller for a Skytron 3500. Do you sell them

RE: Skytron - Elite 3500
Updated  Jul 24 2025
6 Replies

Last year our machine (in a simulation lab) stopped charging and doesn't recognize that it's plugged into an outlet. The power cable and battery have been replaced and the symptoms haven't resolved. I'm suspecting something with the motherboard. Looking at prices I'm trying to determine if it's worth it to go the repair route or recommend replacement. Our CRNA faculty haven't been able to locate a technician to service it. Any advice is appreciated. 

RE: Draeger - Apollo
Updated  Jul 22 2025
1 Reply

How to solve the error sensor defect? In starting time?

Equipment: Stryker pneumo sure high flow insufflator

RE: Stryker - PneumoSure 45L
0 Replies

"Dear Support, I’m an engineer working with the Surtron 400 HP unit, and I would like to know how to access the internal software calibration mode of the device. Could you please provide the procedure or service manual that includes this information? Thank you in advance."

Equipment: LED surtron 400

RE: LED - Surtron 400
Added  Jul 16 2025
0 Replies

Me sale el error

Lámp changer failed en pantalla azul

Y lamp changer failed con vídeo

O vaces parece normal la como si estubiera todo bien pero queda como bloqueado no funciona ni teclado ni nada solo da vídeo

Y También en casos enciende normal  y todo trabaja bien

Me podrían ayudar con esta falla por favor


Translation:

I get the error

Lamp changer failed on blue screen

And lamp changer failed with video

Or sometimes it seems normal as if everything is fine but it remains blocked, neither the keyboard nor anything works, it only gives video

And also in cases it turns on normally and everything works fine

Could you help me with this failure please


RE: Fujinon - EPX 2200
Updated  Jul 12 2025
1 Reply

Tengo un equipo de electrocirugia conmed aspen excalibur

No se como calibrar el equipo Tengo el manual de servició pero falta esa parte de Calibración

Equipment: Conmed aspen excalibur electrocirugia

Translation:

I have a Conmed Aspen Excalibur electrosurgery equipment

I don't know how to calibrate the equipment. I have the service manual but the Calibration part is missing

Equipment: Conmed Aspen Excalibur electrosurgery

RE: Conmed - Aspen Excalibur
Updated  Jul 12 2025
1 Reply

How to change battery

Equipment: Mcquay Alphaclassic Operation Table Battery

RE: MAQUET - AlphaClassic Pro