The densitometer I'm reviewing just had its CSBC card replaced. After that, I tried to run a QA and it always fails in the calibration section of the trend analysis. Could someone help me or share a service manual for this equipment? Equipment: GE Lunar Prodigy
RE: GE HealthCare - Lunar Prodigy AdvanceI would like to ask a troubleshooting guide and possible cause of error code 0x6A ( Generator Fault ) of our existing Stratos Bone Densitometer.
Here is the Error shown in the display.
Here is the equipment Details
Equipment: DMS - Stratos
RE: DMS - StratosLooking for 2nd opinion on a old lunar before replacing detector power supply. The detector test fails. I have photos of many test run
Equipment: GE Lunar bone density
RE: GE HealthCare - Lunar Achilles InSightNeed to repair/service CSBC motherboard,(Rev D).
any advise?
Equipment: GE Lunar Prodigy
RE: GE HealthCare - Lunar Prodigy ProSomeone have the service manual of the GE Prodigy advance?
Equipment: Prodigy Advance
RE: GE HealthCare - Lunar Prodigy AdvanceEquipment: GE progeny Lunar bone density
RE: GE HealthCare - Lunar Prodigy ProHi, my table position does not go to the center when I press the center button. That's why QC has not started.
Equipment: Hologic discovery Wi
RE: Hologic - DiscoveryEquipment: GE HealthCare - Lunar Prodigy
I changed the X-ray block with a second hand one. I aligned the axis of the tube and detector, with original alignment tool. I did the peaking test, that was ok, all values slightly around 2000. I readjusted the collimator too (at signal monitor), where I got around 160000 counts/sec at low energy and 80000 counts-sec at high energy, evenly, at all detectors: - no idea that it is enough or it is a too low value. I made the wobble procedure too.
But after all these at daily QA the Reference Counts fails with around 104.803 cps, 116.237 cps, ratio 0.9.
In the error log I find:
22293: Low energy reference failure
Low energy Reference counts for detector 1# failed. Value 105705
22294: High Energy reference failure
High energy Reference counts for detector 1# failed. Value 118226
22293: Low energy reference failure
Additional detectors fail low energy reference counts
Additional detectors fail high energy reference counts
22294: High Energy reference failure
Anybody any idea where is my mistake?
Equipment: GE HealthCare - Lunar Prodigy
hello, can someone help me with the service tool password ?
thank you
Equipment: GE HealthCare - Lunar Prodigy
Looking for Manual for a Service Manuals For GE Lunar Prodigy we have Encore Version 17.3 on a new Think Center I5 . You can sent it to kdkelley@optonline.net
Equipment: GE lunar prodigy
RE: GE HealthCare - Lunar Prodigy Advance