- All Categories
- General
- Air Cleaners
- Alarm-Occupancy
- Analyzer Measurement & Simulator
- Apheresis Units
- Apps
- Audiometer
- Audiometric Booth
- Battery Testing
- Beds/Stretchers/Mattresses
- Blenders/Regulators/Flowmeters
- Brachytherapy System
- Continuous Positive Airway Pressure Unit (CPA)
- Dental Equipment
- Dictation System
- Electronystagmograph
- Enteral Feeding Pump
- Exam Tables & Chairs
- Facility Boom
- Hand/Foot Switch System
- Handpiece-Dental
- Headwall System
- Hemodialysis Unit
- Humidifier
- Hyperbaric Chamber
- Infant Incubators & Warmers
- Information System
- Infusion Pump
- Irrigation/Distention System
- Laser
- Lensometers
- Lift-Patient Transfer
- Light Source-Fiberoptic
- Lithotripter
- Medical Monitor
- Medication Management System
- Miscellaneous
- Nurse Call System
- Ophthalmic Equipment
- Oxygen Concentrator
- Patient Identification and Security System
- Peritoneal Dialysis Unit
- Phototherapy Unit (Hyperbilirubinemia)
- Point-of-Care Analyzer
- Pulmonary Function Analyzer
- Recorder
- Respirator-Air Purifying
- Scales
- Sequential Compression Device
- Slit Lamp
- Spirometer
- Sterilizers/Autoclaves
- Thermometer
- Traction Table
- Traction Unit
- Urodynamic Measurement System
- Ventilator
- Veterinary Equipment
- Vision Screener
- Warming Unit
- Warming/Cooling Unit
- Washer
- Wheelchairs
- Xenon System
Hemodialysis Unit Forums
Add Hemodialysis Unit to My Bench229 Results
Updated Jul 22 2022
1 ReplyPlease help what to do
C CFF 018 005 BUBBLE CHAMBER (DEAIRATING CHAMBER) ERROR
Unconditional error, blood pump available
General description of conditions for occurring:
- AIVA and REVA opened. Suction pump active and fluid tubes connected
to safety bypass. Low level in bubble chamber is not achieved within 90
sec.
The status from level sensor indicates middle level in chamber.
Technical description of conditions for occurring:
- PRL(P_BubbleTrapHigh)=FALSE
Some possible explanations why the error has occurred:
- The AIVA or REVA could not be opened
- Broken float in bubble chamber
- Broken protective driver board (IC28) RE: Baxter - Gambro AK 96
Added Jul 23 2022
0 RepliesPlease help what to do
C CFF 018 005 BUBBLE CHAMBER (DEAIRATING CHAMBER) ERROR
Unconditional error, blood pump available
General description of conditions for occurring:
- AIVA and REVA opened. Suction pump active and fluid tubes connected
to safety bypass. Low level in bubble chamber is not achieved within 90
sec.
The status from level sensor indicates middle level in chamber.
Technical description of conditions for occurring:
- PRL(P_BubbleTrapHigh)=FALSE
Some possible explanations why the error has occurred:
- The AIVA or REVA could not be opened
- Broken float in bubble chamber
- Broken protective driver board (IC28) RE: Baxter - Gambro AK 96
Added Jul 19 2022
0 RepliesHello,
Error message on NIKKISO DBB-05 asks to do the following checks.
check the connection to the pressure fittings.
Check the PàS2 body installation (blood pump 2)
UP2P mode not selected?
Filter Flooded pressure?
I checked but i don't see the cause.
Please help me. RE: Nikkiso - Model DBB-05
Updated Jun 9 2022
Added Jun 9 2022
Added May 6 2022
0 RepliesHow to do post sample collection for patient. How to off dialysate flow for 20 Sec in 5008s machine RE: Fresenius - 5008/5008s
Added Apr 22 2022
Updated Feb 20 2022
Updated Feb 2 2022
2 RepliesHello, Can any one help by sending me service manual for DBB-07 or how can we change the language?
Thank you! RE: Nikkiso - DBB-07
Added Feb 1 2022