Telemetry-Central Station Forums
Add Telemetry-Central Station to My Bench269 Results
Updated May 14 2020
1 ReplyHello - can anyone tell me how I can obtain a service manual for a Nihon Koden Tele WEP 4208a. RE: Nihon Kohden - Tele-Pro WEP-4208
Updated Apr 13 2020
Updated Mar 18 2020
Updated Mar 6 2020
Updated Feb 4 2020
7 RepliesI want to replace the back cover mainly because there always that one pin get corroded. this pin is for lead set detection. i source out the part, but don't have a procedure to open the case and replace it. any tips? RE: Philips - IntelliVue MX40
Added Jan 13 2020
0 Replies
Per Philips: Proper skin preparation
- Use of quality electrodes
- Change Electrodes Daily
- Proper electrode application
- Good electrode-to-patient contact
- Artifact elimination
If alarming continues despite preparation and/or patient heart status&
- Change arrhythmia analysis mode from multi-lead to single-lead
- Change electrode placement for Electrode LA (black) to V5position on chest
- Change electrode placement for Electrode LL (red) to V6 position on chest
- Change waveform lead to Lead III at Central monitor or Lead that best fits criteria.
Updated Dec 6 2019
8 RepliesHaving an issue here where when SP02 is selected with our telemetry devices and an alarm triggers, when canceling the alarm, it also shuts off SP02 monitoring. This happens with all piics (4) and with all wireless devices (M4841A, and MX40). Any help is greatly appreciated. Thank you. RE: Philips - IntelliVue MX40
Updated Dec 2 2019
2 RepliesHi
I have a Cardiocap/5 which won't display the N2O and anaesthetic agent on the main screen - I know that the machine is able to measure them as somewhere deep in the service menu there was a whole table of all the parameters it was measuring (including N2O and AA). I've had to replace the SRAM battery - have I forgotten to ask for this parameter somewhere?
Thanks RE: Datex Ohmeda - Cardiocap/5
Updated Nov 25 2019
2 RepliesA hospital in Germany donated an M3150 Rev J PIIC Classic to a hospital in Ethiopia I volunteered at. The Sectors and Pt Window is in German, yet the service area is English. The language box is grayed out (shows German). What is the process to change the info the users see into English? The Ethiopian medical community uses English with all of it's equipment. RE: Philips - PIIC iX
Updated Nov 25 2019
4 RepliesI have CMS that i need to integrate with the Philips - IntelliVue MX800,so that the wavelength and parameters can be seen on my CMS. I have been able to set up the ip address on the Philips - IntelliVue MX800 but am not able or cannot see where i can set my CMS ip address on the Monitor.
Is this equipment integratable with other CMS and can i also send the records to PACS RE: Philips - IntelliVue MX800