Follett - REF 2
where to find a follett ref 2 manual

RE: Follett - REF 2
0 Replies

HELMER - iLR125
What PM's are required for Freezers and Refrigerators?

RE: HELMER - iLR125
Updated  Jul 18 2019
4 Replies

Thermo Fisher Scientific - Jewett High-Performance Blood Bank Refrigerators
The hospital wants to repurpose a BB refrigerator for a supply refrigerator, and the temps do not need to be as cold. Adjustment of the controls for the refrigeration system does not apply to the alarm module. Does anyone have a manual to tell me how and if you can adjust the alarm setpoints on the module or if they are locked as-is from the OEM? I know some are locked this way. This one has a "calibration" button which gives me hope that you can do some adjustments on it.

RE: Thermo Fisher Scientific - Jewett High-Performance Blood Bank Refrigerators
Updated  Feb 27 2019
2 Replies

Kirsch - MED-100
The unit is in alarm condition, displaying alternately "Acc" and "5". I don't understand the meaning of this message. Would anybody help me? Any service manual available?

RE: Kirsch - MED-100
Updated  Jan 24 2017
3 Replies

Thermo Fisher Scientific - Forma Blood Bank Refrigerators

Torn Gasket. I need to buy another, for model # 3880. Any ideas??

RE: Thermo Fisher Scientific - Forma Blood Bank Refrigerators
Updated  Sep 7 2016
1 Reply

Does anyone know of a vendor that sells freezer gaskets for ultralow Revco units? Specifically chest freezers.

RE: Refrigerator, Blood Bank
1 Reply

Thermo Fisher Scientific - Revco

I am looking for service manual for the RMI3000S-7-ABA.

This is and old CO2 incubator and I cannot calibrate the CO2 , reading is too high.

I suspect the CO2 sensor. Is anyone got parts for this ?

RE: Thermo Fisher Scientific - Revco
0 Replies

NSBR211 blood bank fridge - does anyone know where on the controller to hook up a two-wire remote alarm on pre-2009 NSBR211 blood bank fridges? The remote monitor needs to see a NO contact that closes under a temperature or power fail alarm condition. The contacts I measure on the controller all read 20Mohms and from 9-14VAC, which puts a voltage back into the monitor node and blows it. I've tried paralleling a second power fail alarm relay off the original one, which worked a couple of times but now the controller won't even give a local alarm when a probe measures past a setpoint.

Any help would be greatly appreciated.

Jeff

RE: Refrigerator, Blood Bank
Updated  Dec 19 2015
1 Reply

Thermo Fisher Scientific - Jewett High-Performance Blood Bank Refrigerators

On the log chart there are two separate incidents of temperature increasing and decreasing over an 8 hr period. Starting at -40 rising to 0 and decreasing back to -40. The unit is a Jewett(now Thermo Scientific)BPL 13 freezer. Any help with this issue would be great.

RE: Thermo Fisher Scientific - Jewett High-Performance Blood Bank Refrigerators

Ask a Question