Patient, Forced-Air Forums

Updated  Oct 10 2016
3 Replies

Checked fuses and power cord. Whats left? replacing board? bair hugger model 505

RE: Arizant - Bair Hugger Model 505
Updated  Aug 19 2016
3 Replies

Is adding 2 oz of propylene glycol a MUST in order to keep the micro temp LT, MODEL 749 disinfected?

What is the purpose of adding propylene glycol?

 

RE: Cincinnati Sub-Zero - Micro-Temp LT
Updated  Jul 28 2016
1 Reply

I have a 505 hose. does anybody need it? I no longer have any 505's.

RE: Arizant - Bair Hugger Model 505
Added  May 11 2015
0 Replies

Do you have the covers for the heating blankets. If so how much are they. We are in need of several,

robinbah@sbcglobal.net

RE: Cincinnati Sub-Zero - Micro-Temp LT
Added  Apr 28 2015
0 Replies

Need a warming element to repair warm touch 5300A.

RE: Nellcor - Warm Touch 5300A
0 Replies

Arizant Bear Paws Service Procedures

PM Process:

  1. Perform electrical safety procedure.
  2. Install temperature test apparatus on end of hose.
  3. Expose test pots by lowering filter shroud.
  4. Replace filter.
  5. Extend hose to full length.
  6. Turn unit on to full value of thermostat control.
  7. Allow 10 minutes to stabilize
  8. Take reading (43.0C +/- 0.5)(if out, see calibration below).
  9. Push over-temp switch to advance temperature uncontrolled.
  10. Alarm lamp should light below 56C (There is no control of this value).

Calibration Process:

  1. Perform PM procedure until step 6 completed.
  2. Using small flat screwdriver, adjust temp reading gradualy through adjust hole under shroud.
  3. When 43.0C is met and stabilized, proceed with over-temp steps 7 & 8 above. 

RE: Arizant - Bair Paws Model 875
2 Replies

I have been looking for a part number for the air filter for these units with no luck. It is not in the service or operator manual for some strange reason. They are owned by 3M now, and neither set of manuals have this part number.. weird... Thanks if you can help me with a part number.

RE: Arizant - Bair Paws Model 875
Added  Dec 2 2014
0 Replies

STABNDBY LED AND 46 TEMPERATURE CONTROL BUTTONS ARE WORKING.

RE: Care Essentials - Cocoon CWS 4000
Added  Nov 19 2014
0 Replies

Can you use Diversey Virex II 256 or Oxivir TB wipes to disinfect and clean this unit?

RE: Gaymar - Medi Therm MTA-4700
Added  Oct 22 2014
0 Replies

where can I find a replacement hose. Are there any specific for this hose.

Thanks.

RE: Nellcor - Warm Touch 5300A