Exam Tables & Chairs Forums

6 Replies

This morning the ultrasound tech completed for study without issue. Bed elevated/lowered without issue.

When second patient on table, the bed will not change position in response to selection of any buttons on the control. We hear a beep each time make a selection. 

Have unplugged bed for more than one minute. Plugged back in . Still not working. 

All cords seen in good working order and securely in place. 

The ultrasound table is an "Oakworks general with fowler"

Thoughts/Suggestions?

RE: Oakworks - General w/ Fowler
1 Reply

Customer’s unit stuck in elevated position with no articulation. 

I’m thinking could be a bad MCB but need to check all the limit switches first. 

Does anyone have a service manual for this exam chair? TY!

Email to: James@TurnerMedicalRepair.com

RE: Midmark - 417
Updated  Jan 3 2024
4 Replies

Customer has a 630-020. Chair has up/down motion with foot and hand control but no other articulation. 

Need advice on probable causes and service manual. 

TY! James@TurnerMedicalRepair.com

RE: Midmark - Ritter 630
Updated  Nov 13 2023
3 Replies

The end shelf was bent by a patient , is it possible to replace?

Equipment: Ritter 204 exam bed

RE: Midmark - Ritter 204
Updated  Nov 3 2023
1 Reply

parts availability

Equipment: midmark 411

RE: Midmark - 411
Updated  Oct 16 2023
2 Replies

The chair drifts backwards even though the recline button is not depressed, especially for heavier patients.

Equipment: Midmark - 75L

RE: Midmark - 75L
Updated  Aug 31 2023
2 Replies

Have a Ritter 223 where chair will raise and lower but back rest is in lowered position and will not raise. Customer has new hand control. Switched back out to old hand control still nothing. No noticeable noises or anything when hand control activated. Any leads on where to start? Clinic is an hour away and want to be prepared. Thanks guys!

RE: Midmark - Ritter 223
Updated  Aug 11 2023
2 Replies

Can I clean the mattress with "Healthcare Clorox Bleach Germicidal wipes"?

RE: Durabuilt Medical - PMT 8000 HLT
2 Replies

The remote moves all the parts of the chair except for the foot section.  There is power and you can hear it wanting to move but nothing happens.  Checked the cords and power to that part of the table.  All looks good.

RE: UMF - 5020 Power Procedure Table
Updated  Jul 19 2023
6 Replies

 NO BUTTONS WORK AND MOTOR IS RUNNING  (IS NOW IN HIGHEST UP POSITION) UNPLUGGED-

ST CHARLES ILLINOIS 

RE: Midmark - 117