- All Categories
- Laboratory
- Amino Acid Analyzer
- Automation System, Laboratory
- Bilirubinometer
- Biomedical Freezer
- Biotech/Life Science
- Blood Culture Analyzer
- Blood Gas/pH Analyzer
- Blood Grouping System
- Cell Processor
- Centrifuge
- Clinical Chemistry Analyzer
- Coagulation Analyzer
- Cryostat
- Cytometer, Flow
- Electrolyte Analyzers
- Electrolyte Analyzers
- Electrophoresis System
- Embedding Center
- Glycohemoglobin Analyzer
- Hematology Analyzer
- Histology Analyzer
- Histology Printer
- Immunoassay Systems
- Incubator, Laboratory
- Laminar Airflow Hood
- Microbiological System
- Microplate
- Microscopy
- Microtome
- Oximeter, In Vitro, Laboratory
- Photometric Microplate Readers
- Platelet Aggregation Analyzer
- Refractometer
- Refrigerator, Blood Bank
- Slide Stainer/Prep
- Spectrometer, Laboratory
- Spectrophotometer
- Thyroid Uptake System
- Tissue Processor
- Ultrasonic Cleaners
- Urine Analyzers
- Warming Cabinet
- Washer/Decontaminator
Cryostat Forums
Add Cryostat to My Bench176 Results
Added Jun 28 2022
0 RepliesHello I'm looking for the service manual for an HM 550 or if not HM560.
Thank you RE: Richard-Allan Scientific - Microm HM560 CryoStar
Added Jun 22 2022
Added Apr 8 2022
0 RepliesDos nyone know what is the exact replacement for the HM525 controller circuit board? Is the BERGER LAHR CONTROL CIRCUIT BOARD CARD LS 220642-00-01 C LS 2206420001C the correct replacment? I have seen it listed as a replacement for the HM505 but was not sure whether it is the same for the HM525. RE: Microm - HM520/HM525
Added Apr 7 2022
0 Replies Specimen head temperature doesn't work
I changed the Peltier object and checked in the service mode how much current it draws
only 1600mA comparing to the original one from anther cryostat x50 that draw 2400mA
and the temperature reaches only -3 in the best cases.
so is the problem the new peltries because i,ve changed twice and still the problem occurs
please help
thanks
Mohannad
RE: Thermo Fisher Scientific - NX50 Cryostat
Updated Mar 24 2022
1 ReplyThermo Shandon Cryotome E (620E).
Does anyone know what the P5 error code indicates? It was in the micron setting window. I'm not sure it's an error, but I don't know what it indicates. RE: Thermo Fisher Scientific - Cryotome E
Updated Mar 16 2022
1 Reply
Cryostat not advancing. Best way to check. First time on Sakura. Have always worked on Leica
Updated Dec 14 2021
Updated Nov 10 2021
2 Repliesneed help decoding error EO: 15 (Leica Biosystems CM1850 UV) RE: Leica Biosystems - CM1850 UV
Updated Nov 9 2021
2 RepliesHas anyone had any luck reseting the HELP message on the display of this cryostat? RE: Leica Biosystems - CM1860
Added Oct 15 2021
0 RepliesLeica CM1520 cryostat.
I am looking for the current part number for the expansion valve module. I was told that the e-valve module from the CM1850 would work, but Leica says that both part numbers I have for that have been discontinued, with no replacement.
I have been unable to find a service manual or parts list for the 1520.
RE: Leica Biosystems - CM1520