To start at the beginning ... 1st) Tube/collimator was tough to move up/down to column. Replaced upper brake to column, brake wires shorted to housing when re-assembling. Repaired wires, routed wires a different way, re-assembled ... Tube/collimator easy to move up/down. Charged portable overnight, next day went to verify system working before giving to the techs .... rotor cuts out in 2 seconds & system restarts, tests passed, no errors, CB never trips off.
2nd) Now AMX-4 + when you rotor from Hand switch, can hear relays click, 2 seconds later the displays go blank and startup test pass, display is ready again. I can hear the rotor spinning in tube, tube resistance is good. I verified capacitor values on right side. All fuses are good. Reseated Relays on circuit boards. I have a replaced 1 board at time to the following .... Display Bd, CPU, Rotor Bd, FIL/Kvp Bd. No errors appear after startup tests. Can't do any cals because rotor doesn't last but 2 seconds. Does anyone have any other idea's?
Thanks
Jeremiah
Equipment: GE HealthCare - AMX 4 Plus
RE: GE HealthCare - AMX 4 PlusEquipment: Shimadzu - MobileArt Evolution MX7
Model 46-270954G1 AMX4 July 1999
When lock release is pressed at collimator handles, vertical nor transvers locks will release. Both locks and switches check good. I do not get the voltage on the lock wires to disengage them. It looks like there is more to the circuit than power, switch, and locks. Anyone have wiring diagram for this model? or any advice on repair. Thanks
Equipment: GE HealthCare - AMX 4
Equipment: Amx4
RE: GE HealthCare - AMX 4What does that error indicate? 70100045
Equipment: Carestream - DRX-Revolution
RE: Carestream - DRX-RevolutionI have facing problem when exposure button is pressed error papers 431 & 433 any help to resolvsolve this
RE: GE HealthCare - AMX 4 PlusHello, I have a big problem, I have this portable x-ray that gives me code d13, I changed the button that releases the brake and the same problem continues, please I need help
Equipment: Shimadzu MUX-200D-XC
RE: Shimadzu - MobileArt Plus MUX-100HThe charging process stops after few minutes . please suggest a solution.
RE: Shimadzu - MobileArt Evolution MX7