0 RepliesDoes anyone know what this error code means. The batteries are not charging also. RE: Fujifilm - FDR Go
Added Mar 23 2023
0 Replieshi i need the service password for xray optima xr 646
i know the user is Ge Service,
also i try with insite, 2getin , but doesnt work
best regards
Equipment: ge optima xr646
RE: GE Healthcare - Optima XR646Added Mar 20 2023
0 Replies- The connection with the generator cannot be established
SPMMESS12
- generator error 31 occurred. (Ret SPMMESS41
- Not possible to set generator params, please retry Ref
SPMMESS221 RE: AGFA - DR 400
Added Mar 10 2023
0 RepliesWhat is the most common fix for error code 40480/80? Performed invertor calibrations with out issue. Intermittent early termination of exposure
Equipment: Multix Select DR
RE: Siemens - Multix Select DRAdded Mar 5 2023
0 RepliesI am looking for an install manual for this unit.
I would appreciate if anyone could help me out with this.
Thank you RE: AGFA - DR 400
Added Feb 24 2023
0 RepliesNew installation (unit moved) I cannot register cassettes. illegal barcode message . Do i need barcode reader?
Equipment: DRGEM - GXR Series
RE: DRGEM - GXR SeriesAdded Feb 23 2023
Added Feb 15 2023
0 Replies1. J Error 13 38 V DC supply out of range
Check the value with a volt meter on the test point of the Power Supply
Board and replace it if the value is not in the range 34 to 42 V.
Voltages are Ok , I have measured it and its 36Volts. This error seldom occurs...
Q Error 21 No tube current
No or low anode current. This error can be caused
• by interruption of High Voltage power supply (right side of the Power
Switching Unit board)
• or by interruption of the filament power supply (left side of the Power
Switching Unit board)
• or by interruption of the driving circuits (in the regulation board)
• or by fault in the tube-head (input transformer or multiplying circuits).
Check the supplying circuit in the bushing (multi-pole cable) and the flat
cable between the regulation and power switching boards.
On the Power Supply unit check the switch SA 301 (must be ON), fuses
FU 304 and FU 305. If they are OK, switch exposure and check the driv-
ing signals for HV regulator and inverter (points 1, 2, 6 and 7 on the Regu-
lation Unit) and the HV inverter supplying voltage (between points D3
and S5 in the Power Switching Unit) – should be in range from 50 to 100
volts (depending on kV and mA setting). Then measure the primary circuit
of HV transformer.
If all is correct, the circuit between the secondary of HV transformer and
tube is broken.
Depending on the location of the fault, consider to replace the Power
Switching Unit board, the Regulation Unit board, the connecting cables or
the tube-head.
This error can be also activated by sparking in the tube, when a very short
over-current signal switches off the HV regulator only, but is not trapped
by the Control Unit in order to activate error 17.
In this case try to perform the tube seasoning (de-gassing by making a
number of exposures at low kV) or replace the tube-head if necessary.
This is difficult to find. I am currently trying to check the cabling.... Any ideas further? RE: BIOTEX S.A. - BlueX PantOs DG XP
Updated Feb 15 2023
3 RepliesI found an error 44 for my machine i need i help on how to solve it
Equipment: DRGEM - GXR Series
RE: DRGEM - GXR Series