STERIS - Amsco 3052
STERIS Amsco 3052 Forums
Added Dec 20 2019
0 RepliesDuring Instrument cycle on Wash 2, chemical used was Detergent. While injecting, washer alarmed "16. Chemical Injected Volume Too Low." The alarm would pop up in the middle of the day, which made me believe that it wasn't a check valve issue next to float sensor for the chemicals. I replaced all hoses for the Knight peristaltic pumps and added silicone grease. I disassembled all valves and flow sensors to check its condition (all seemed in good condition), so I was thinking that it would be a faulty intermittent flow sensor on detergent valve. I also calibrated volume delivery accuracy for all pumps, and yet it failed with the same alarm. I recently replaced detergent valve with an upgraded kit (Chemical Valve Installation PN: 10094478). It ran fine for a week and now the error came back.
Has anyone ever encountered the same issue? Any suggestion would help! RE: STERIS - Amsco 3052
Updated Aug 16 2019
1 ReplyDoes anyone have a service manual they can email me please! My email is tkussmaul@centeronesc.com, Thanks!!! RE: STERIS - Amsco 3052