Hologic - Lorad M-IV Series
by Hologic

Hologic Lorad M-IV Series Forums

0 Replies

Hello how do i log in to the root user of the Selenia Hologic mammography I bought it a second hand so the router is not wrong hence I can share images 

Equipment: Hologic - Lorad M-IV Series

RE: Hologic - Lorad M-IV Series
Updated  Jul 24 2024
2 Replies

Hello,


We have a Hologic Selenia here.

It's giving us an error.

The error code is M-IV Error: 78

I saw on another post that it is the mag tray detect board. I cleaned it however it is still not working. Is this the correct part that is having issues?

I don't see error 78 on any service manual.

RE: Hologic - Lorad M-IV Series
Added  Jul 28 2023
0 Replies

MANUAL SERVICE

Equipment: Hologic - Lorad M-IV Series

RE: Hologic - Lorad M-IV Series
Updated  Jul 3 2023
1 Reply

Hi dear forum. I am seeking help in resolving a problem with a Hologic Lorad Selenia.  Can help me with this error ? "Error during worklist query" ,how to solve this problem?
I hope that someone can give me support for solving this problem. Thanks in advance. Regards.

RE: Hologic - Lorad M-IV Series
Updated  Apr 27 2023
1 Reply

Hello colleagues I would really Appreciate if you can help me with this error 

RE: Hologic - Lorad M-IV Series
Added  Apr 27 2023
0 Replies

Hello colleagues I would really Appreciate if you can help me with this error 

RE: Hologic - Lorad M-IV Series
Added  Apr 27 2023
0 Replies

Hello colleagues I would really Appreciate if you can help me with this error 

RE: Hologic - Lorad M-IV Series
Updated  Mar 24 2023
1 Reply

Hi friends I am a biomedical engineer here in Mexico, and i got this equipment. The problem is that when my technicians want to use the auto-modes, sometimes they got this error: "Calculated Exposure time exceed back up time" . I dont know it has to do something with a bad technique. Also the images when that happens, the breast is totally white. Manually i do not have that problem.

RE: Hologic - Lorad M-IV Series
Updated  Oct 12 2022
1 Reply

MIV Error 28

Equipment: Hologic - Lorad M-IV Series

RE: Hologic - Lorad M-IV Series
Added  Dec 27 2021
0 Replies

Good morning dear forum. I am seeking help in resolving a problem with a Lorad MIV mammography system. The problem is that there is too much scattered radiation outside the collimated area. As a test, I completely close the collimator and still cannot remove the scattered radiation. Could someone give me any suggestions to solve this problem? Any help will be highly appreciated. Thanks in advance. My e-mail: vicmaal@yahoo.com.mx Greetings.

RE: Hologic - Lorad M-IV Series