1 ReplyI have an S222-HD table that won't go down. The lift motor is fine; if I wire it backwards it'll go down fine. It isn't the obstacle switches or the limit switches.
The control circuit board has been replaced twice. In either case, when I read voltage across the motor terminals on the control board while pressing the "down" switch, I see 0.365VDC, so something is coming across the relays, just not the 24VDC that should be crossing.
Summit won't give any tech support, neither will NHD, who has their name all over it.
Anyone out there have a schematic or an answer? The "service/installation manual" on site from Summit doesn't have any schematics, theory of operation, or troubleshooting.
I'm at my wit's end here . . . RE: Summit Industries - Radiology Room Systems
Added Mar 30 2021
0 RepliesWhat is the table weight limit for a summit s222-hd xray table? RE: Summit Industries - Radiology Room Systems
Updated Feb 5 2020
1 ReplyHi, I have a summit 325e that the red tube light is on and it will not shoot x-rays regardless of the exposure settings. The rotor comes up to speed, but the contractor does not close. Any help would be appreciated. I am currently at a small hospital in rural Haiti