Philips - HD 7
Philips HD 7 Forums
Hi all,
I have a Philips HD7 XE ultrasound system with the following specs:
- Part Number: 989605368681
- Serial Number: CI51090020
- Hardware Rev: 2.0
- Software version: 454110266224__2.0__HD7_Ultrasound_Application__[2009/5/06]
- Manufactured: June 2009
The system fails to boot with BSOD:
Disk (Seagate Barracuda ST3250310AS) has been tested — SMART passed, extended self-test passed, no bad sectors, no I/O errors. The issue is purely the SOFTWARE registry hive content.
Attempted: ntfsfix (filesystem was unclean, now fixed), chntpw inspection (hive partially navigable but corrupt), truncating SOFTWARE.LOG. No restore points, no repair folder on this XP Embedded installation.
I need either:
- A disk image for this exact system (HD7 XE, HW Rev 2.0, SW 2.0)
- The SOFTWARE registry hive file from a working identical system
- Any known repair procedure specific to this platform
Philips support confirmed they cannot assist — system is out of support.
Any help appreciated.
Equipment: Philips HD7
RE: Philips - HD 7The manual says that when you press select a cursor appears, and pressing select again another cursor appears and now pressing select button now toggles control between two cursors so how do you make a measurement and save it? thanks
RE: Philips - HD 7Hi!
Would you be so mkind to send me info about the service manual for HD7?
Thanks in advance,
Marko
RE: Philips - HD 7