Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Bio-Rad Laboratories, Inc. C1000 Touch Thermal Cycler Forums
Add Biotech/Life Science to My Bench8 Results
Updated Mar 9 2023
2 RepliesThe C1000 touch passed system start up tests. Ran the diagnostic test under administrator and passed. I accidentally started a second diagnostic test and was trying to cancel out and I clicked no to the lid not shutting. After the system would not pass system start up test. There was an error already present but not affecting function (x000D000D) slow cycling lid. Thanks for any insight. RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Added Mar 27 2022
0 RepliesOlá estou com um termociclador c1000 apresentando erro 29 POWER BLOCK FAILURE A e B. Testei transistores e capacitores dos circuitos das fontes e aparentam ok. Fiz a troca como sugerido em outros posts dos caps de 100uFx25V próximos ao U11 UC2844, sem resultado. Observei que o transistor Q9 IRFB3030 aquece bastante e ele não tem dissipador, mas esta ok nas medidas, e mais nenhum outro componente aquecendo Outra característica que esse defeito esta apresentando, ao ligar a chave on se escuta algumas oscilações. Alguém ja presenciou esse problema? Muito obrigado pela ajuda. RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Updated Mar 16 2022
8 RepliesI am trying to figure out where block 5 is. I pulled circuit boards and tested all temp sensors. Sensors tested good. Lid sensors and resistors tested in spec.. As I have no schematics (not available) I cannot trace any circuits easily to find TEM drivers and other possible suspect components.
Any ideas would be welcome!
Thanks,
Dan RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Updated Feb 10 2022
11 RepliesHas anyone encountered the error "Self-Test Failed on X' on BioRad C1000 touch.
Please help me out on how to solve it RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Updated Sep 30 2021
1 ReplyIs part # 10013441 the correct Real-Time Alpha Block? RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Updated Apr 30 2021
4 RepliesI have a C1000 which won't power up. I found F3, a 5x20mm 5A fast fuse, blown. After replacing it, F103, a 5x20mm 500mA fast fuse hidden under the power inlet, blew when I applied power, but before I turned on the power switch.
RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Added Jan 29 2021
0 RepliesHello, BIO-RAD C1000 during sturtup procedure fails with two errors:
A[26] Lid overheated and was shut off. Please
call Bio-Rad for service.
B[26] Lid overheated and was shut off. Please
call Bio-Rad for service.
Maybe you can provide more information about this error from service manual?
Thank you. RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler
Updated Jul 12 2020
21 RepliesThe BioRad C1000 Thermal Cycler makes clicking sounds on startup. Sounds like a relay. Could somebody guide me on finding the problem? Is there a service manual or a schematics of this device? RE: Bio-Rad Laboratories, Inc. - C1000 Touch Thermal Cycler