Does anyone have any tips / tricks to adjust the delrin block? I'm struggling to get even pressure across the travel. Any thoughts / suggestions would be greatly appreciated.
RE: Leica Microsystems - RM2125 RTI am in need of a service manual for the Leica RM2125RT. Can anyone help?
RE: Leica Microsystems - RM2125 RTThose of you who refurbish the Leica RM2125 or 2125RT (also the Sakura SMR 200), do you routinely (or often) replace any parts in the micrometer mechanism?
Ross
RE: Leica Microsystems - RM2125 RTA few histo-techs like really tight rotary microtomes - high turning force. I have one where I keep adjusting the bar above the link block to yield the high turning force the tech prefers (three times the spec), but it doesn't stay that way long enough. Perhaps the bar creeps (despite tight screws), or the block wears, or both. I was wondering if a heavier grease might be part of the solution. Has anyone else addressed this issue?
RE: Leica Microsystems - RM2125 RT