STERIS - Reliance Vision Single Chamber
by STERIS

STERIS Reliance Vision Single Chamber Forums

0 Replies

Every so often there is water dripping off the clean side control panel. I know the water is dripping from the HEPA filter assembly above the panel. The HEPA filter and prefilter is dry. The condensation system is working. I have replaced all the worn plastic pieces that might be spraying water in the direction of the vent. The blower is turning in the correct direction. There is no leaking fluid tubing over it. It doesn't do it on every cycle ran or on a specific cycle. My question is, How is the water getting in the HEPA assembly filter if air is drawn through it from the outside? All the checks mentioned above were recommended by Steris Tech Support.

RE: STERIS - Reliance Vision Single Chamber
Updated  Nov 19 2020
1 Reply

Halfway through a cycle, all actions stop operating. None of the valves turned on when they were supposed to...doors wouldn't open. I went into service mode to inputs/outputs and tested the door and all the water valves and nothing would operate/energize. I checked the incoming air at the regulator and it was fine. What else should I be checking? What could cause this? Thank you.

RE: STERIS - Reliance Vision Single Chamber
0 Replies

At startup, the system goes to the desktop screen then the Steris screen with the progress bar. After the progress bar, it still just sits on the Steris screen. Can someone tell me how to access the Softlogix Controller to verify it's on "run'' mode? Thank you.

RE: STERIS - Reliance Vision Single Chamber
Added  May 11 2020
0 Replies

CPU and hard drive went out. Replaced Allen Bradley CPU and cloned hard drive from another SC Vision Washer. Turned unit on and it went thru its normal booting process, then it stopped and asked for the Administrator Password. called Steris and they said it was an AB issue, called AB and they said it was a Steris issue. Sounds almost like Steris doesn't want to help because I didn't purchase the CPU from them.

RE: STERIS - Reliance Vision Single Chamber
1 Reply

trying to get into the Service mode to change some settings.

RE: STERIS - Reliance Vision Single Chamber
Updated  May 24 2019
4 Replies

Hello, I am having a very strange issue with one of my Vision washers. Twice tonight I have gotten calls with the customer stating that the chamber is overfilling and having nearly two feet of water in it. I have tested the water level sensors in service mode and through running cycles. They are working fine. The pneumatics are opening the drain valve fine, drain line doesn't seem to be clogged, drain filter isn't dirty, water inlet valves aren't leaking. Everything checks out. However, both times that I have come in, the rack in the chamber has been empty, and the water has been completely clear. No chemicals or anything. So basically I'm wondering if there is any possible way of this happening without someone overfilling the chamber in service mode, as some of our staff unfortunately know the service passwords. Any help would be greatly appreciated, thank you.

RE: STERIS - Reliance Vision Single Chamber
Updated  Aug 14 2018
3 Replies

Hello, I am looking for the water consumption of this instrument as well as the recommended PSI.

RE: STERIS - Reliance Vision Single Chamber
Updated  Oct 15 2015
2 Replies

The door open/closed sensor on the load side doesn't sense the door is closed and will not allow a cycle to be started. Is there a way to bypass this so the washer can be used while we're waiting for repair?

RE: STERIS - Reliance Vision Single Chamber
Updated  Dec 10 2014
1 Reply

Looking for a service and operator's manual for a Vision Washer

RE: STERIS - Reliance Vision Single Chamber

Start a Thread

Equipment Name: STERIS - Reliance Vision Single Chamber
SIGN IN FIRST
This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree