Sorvall - Legend T/RT Plus
Sorvall Legend T/RT Plus Forums
Hi
our machine has stopped working and displays the error e30
after looking through the manual, it says replace pcb, unfortunately, they are no longer made and the only ones i can find are in America from used machines.
i have 2 questions
1, do you think its worth sending the pcb off to a company that repairs pcbs to see if it can be fixed (i have asked a few near me but they dont work on centrifuges)
2, if so, is the pcb the same as the main board, on the schematics, it looks like it is but in the instructions, it refers to main board and pcb seperately.
many thanks and i appreciate any help
RE: Sorvall - Legend T/RT PlusThe seven segment displays either have an "8" or a small "L" (upper left and middle segments), 1, 2, 3, 4, and the snowflake all have yellow lights. The unit has been powered off and on again. The manual mentions "display remains dark" and "display elements are shortly illuminated." I don't think either quite fits. Thoughts?
RE: Sorvall - Legend T/RT PlusI can't get the lid to close- any suggestions, experience or ides?
RE: Sorvall - Legend T/RT PlusWe have the same problem with the error code e-41 on our sorvall legend. I have tried doing what was posted on the discussion like checking the ribbon from the motor attached on the PBC board.
Clarification about the ribbon cable, is it the ribbon cable fron the motor going to the main board or you are referring to other ribbon cable?
Well i did check all the cables running on the main board and see to it that they are all well attached, and so they are. After doing that i tested it again and it is working fine again. The next week after testing it during the weekends, when i turned it "ON" again, the ERROR CODE E-41 appears again.
Is there any other useful information that anyone can share about this ERROR CDE E - 41? Any other recommendation or parts to be replaced to solve this problem? Please HELP
RE: Sorvall - Legend T/RT PlusWe have the same problem with the error code e-41 on our sorvall legend. I have tried doing what was posted on the discussion like checking the ribbon from the motor attached on the PBC board.
Clarification about the ribbon cable, is it the ribbon cable fron the motor going to the main board or you are referring to other ribbon cable?
Well i did check all the cables running on the main board and see to it that they are all well attached, and so they are. After doing that i tested it again and it is working fine again. The next week after testing it during the weekends, when i turned it "ON" again, the ERROR CODE E-41 appears again.
Is there any other useful information that anyone can share about this ERROR CDE E - 41? Any other recommendation or parts to be replaced to solve this problem? Please HELPPPPPPPPPPPP
RE: Sorvall - Legend T/RT Plus