I would liké to know thé Caractéristique of thé lampe ? Is it 12v 20watt. ?

RE: Hospitex Diagnostics - Mini Screen p
Added  Dec 31 2015
0 Replies

i need to change the lamp

can i use lamp 12 v 50 watt

RE: Hospitex Diagnostics - Mini Screen p
0 Replies

The unit could not be measured the Enzyme Tests such as GOT, GPT (340 nm filter) and Alkaline Phosphate (405 nm filter), but another tests can be measure as well.

In Filter Tension Check, I already followed the procedure as well and I can regulate the Trimmer P1 to obtain the (OPT.TENS) about 300 mV in 340 nm Filter, however, the OPT.TENS of the 340 nm Filter is not stable whenever I switch ON and then checked it again, it varies from 100 mV to 300mV. Sometimes it shows about 180 mV and then stable at near this point, but sometimes it shows about 250 mV and then stable at near this point, sometimes it shows other values.

But some of these Instruments are stable about 300 mV in 340 nm Filter. However all of these Instruments cannot be measured the Enzyme Tests

RE: Hospitex Diagnostics - Mini Screen p