I am facing the communication errors 0075 "Bus-Comp hardware failure" and 0077 "Bus-Coll hardware failure". It seems like a common failure, so I checked the communication cables: between the supervisor and Stereo PCBs, the cable between the Stereo and Collimator PCBs, the cable between the Stereo and Compressor PCBs, all three cables gave good continuity. The communication indicator LEDs (RXD, TXD) on the supervisor, Stereo and Compression PCBs are dimly lit, but on the Collimator PCB the LEDs are off. It appears that the Collimator PCB is bad, but the unit indicates two errors from two different devices (Collimator and Compression) as if it were a common problem on the Stereo or Supervisor cards. I really appreciate the help on this. Greetings.
RE: GE HealthCare - DiamondHi.
Every one I need GE diamond memoghraphy service manual
Any one have service manual
Please send me
Equipment: GE Diamond Memoghraphy
RE: GE HealthCare - DiamondI have a Dimond Mammo that has an intermittant error of er0087 and it locks up the cassete. You and hitok and it continues as normal. Does anyone know what the problem is?
Thank you!
RE: GE HealthCare - Diamondhi , I need the Diamond mammography service software, please contact mauro@garudamedical.com.br