Hi, the system has intermittent E01 generator Failure.
When it happens, I check prep is fine and rotor spinning sounds smooth. But there is no exposure at all.
As soon as press exposure button, it tries to make exposure, but it fails and shows the error on console maybe in 3 sec.
I measure input voltage is 204VAC and 200VAC is selected on Mains Distribution board.
I also check LEDs on all boards in the side cabinet, and all seems fine.
Please share any of ideas to deal with E01 Generator Failure.
RE: GE HealthCare - Senographe EssentialHi,
I am a medical physicist and i have to check the quality of an Essential mammograph. I am not expert in DR mammography so i would like to know your opinion about some quality control results, especially about uniformity and artifact evaluation (variance map). May I send someone some results or can someone send me some uniformity and variance map images (with adjusted brightness and contrast)?
(My email: fra@quotazero.com)
Thank you!
Francesco
RE: GE HealthCare - Senographe EssentialEquipment: GE Healthcare - Senographe Essential