Anybody has a calibration password for Ge Essential Mammo? Thanks
RE: GE HealthCare - Senographe EssentialIm getting E10 Heating failure track 1 error. I can only expose with Rh target, The Mo target is not available therefore, I cant perform the flat field test. Any suggestions?
RE: GE HealthCare - Senographe EssentialDoes anyone know if the footswitches are supposed to be deactivated at compression of 3 daN and over? The keypad buttons on the sides have no problem.
RE: GE HealthCare - Senographe Essential
Hello,
I have a GE Essential that is getting error S60- X-ray beam quality error- x-ray are disables and it won't let you expose. Could it be my tube? I was going through gain calibration when this error occurred- I was 2/3 of the way complete- starting the Mag gain calibrations when this happened. All previous calibrations were fine.
RE: GE HealthCare - Senographe Essential