I have a Midmark M11 Sterilizer that is maintained weekly and monthly according to manufacturers instructions. The last two times during the weekly maintenance, we have notice tiny dark colored specks in the drain water. What could this be? The element and filters appear to be in good shape. Will this affect the effectiveness of the sterilizer?
RE: Midmark - Ritter M11GOOD AFTERNOON MY PROBLEM IS ERROR E 010 OR E011. I WAA LOOKING FOR A HARDWARE PROBLEM BUT I DID NOT FIND IT. MAYBE, IS IT THE MEMORY OF THE BOARD? THANK YOU SO MUCH.
where can I get the board? year built 1996
RE: Midmark - Ritter M11Refer to operator’s manual for detailed instructions.
Weekly
1. Clean external and internal surfaces
2. Change water in reservoir with clean, distilled water.
Monthly
1. Flush system
-Drain and fill reservoir with clean distilled water.
-Add one ounce of SPEED-CLEAN to cool chamber.
-Run one cycle at 270?F for six minutes. Instruments should not be sterilized during cleaning.
-Drain reservoir and refill with clean distilled water.
-Run one unwrapped cycle.
-Drain reservoir and after cooling clean trays, tray plate, rack, and filter.(refer to operators manual for details)
-Refill with clean, distilled water.
2.Check pressure relief valve.