Hologic - QDR 4500
Hologic QDR 4500 Forums
Add Bone Densitometer to My Bench26 Results
Updated Aug 15 2023
4 Replies
Over the weekend, I replaced the X belt. Upon completing the installation, we discovered that the table centering wasn't operational; it only moves in one direction. The daily check also fails, and when the phantom is placed, the table doesn't center and displays error 27. I attempted motor TX calibration, which saved the values, but the issue persists. Is there anyone who can assist me in resolving this?
RE: Hologic - QDR 4500Updated Mar 14 2023
1 ReplyState machine error task 0 index 263 what is mean and from were hardware or software?
Equipment: Hologic - QDR 4500
RE: Hologic - QDR 4500Added Jun 17 2022
Updated Dec 18 2020
Updated Aug 26 2020
6 RepliesI am getting an error when doing my daily QC, ' Exist, repeat the QC procedure and check the phantom position carefully. System test failed. Check that the phantom is the only object in the table.'
Position of phantom is in the right spot as the case in everyday daily QC.
Any ideas how to proceed?
Thank you RE: Hologic - QDR 4500
Updated Jun 18 2019
2 RepliesI´m a printing service provider and I need to replace the computer where the software are installed because I need to print in a usb printer, i don´t have the QDR software and divers, can you help me?, thanks and best regards RE: Hologic - QDR 4500
Added Aug 27 2018
0 RepliesHello guys, Anybody knows how to change the Ethnicity in HOLOGIC QDR 4500 BMD..Please let me advise if u anyone know,
Thankyou
syed RE: Hologic - QDR 4500
Updated Dec 5 2017
Updated Sep 6 2017
Added Jun 2 2017
0 RepliesError 9173 IDS_ERR_SCN_ENDSCAN logged (Stepper motor thermal overload circuit trips). ? I cleaned long. motor tracks and inspected drive belt ok. Ran excursions could not reproduce. Possible faulty stepper motor ? RE: Hologic - QDR 4500