CareFusion - Alaris MedSystem III 2863
by CareFusion

CareFusion Alaris MedSystem III 2863 Forums

1 Reply

I have a MSIII model 2863 with a customer complaining of short battery life. The device charge current is 180mA, which is in spec according to the manual. Batteries charged 24 hours in the device have a very low capacity. Testing the batteries in a Cadex analyzer with charge current set to 180mA, and the batteries never reach full charge. Increasing charge current to 500mA increases battery capacity to 100%. I've found this to be the case on numerous batteries, and they are all R&D. Any ideas?

RE: CareFusion - Alaris MedSystem III 2863
4 Replies

What are the host system requirements for the MED SYstem III Field Maintenance Software that is used to support the MEDSYSTEM III Infusion Pumps?