ATL - HDI 3000
ATL HDI 3000 Forums
Add Ultrasound Systems to My Bench50 Results
Updated Aug 18 2023
Added Dec 7 2021
0 RepliesHola a todos. Ultrasonido ATL HDI 3000.
I need help, this equipment was working properly, at this time when I turn it on at the end of the boot process it does not show me the sweep and does not allow me to select any transducer, when I select the SCAN HEAD key I get the screen to select the transducers but practically without information, 4 boxes appear above without any information inside and below another box appears with the name Tissue Specific Presets but also without any information. Please someone can tell me to refer this problem to and how I can solve it. Thanks a lot.
I measured the voltages of the sources ps1, ps2 and ps3, note that ps2 and ps3 are present voltages, in ps1 +/- 15vdc and +/- 6vdc, in tp8 I have 155 vdc and in tp1 and tp2 I have +/- 31vdc. Will this be correct?
thank you RE: ATL - HDI 3000
Updated Nov 4 2021
Updated Mar 25 2021
1 ReplyAfter replacing monitor upon boot up I am getting a yellow alert. That I can't remove from the screen. I checked but no current support techs near me.
RE: ATL - HDI 3000
Updated Jun 25 2020
2 Repliesi'd like to use 15 pin monitor how do i go about the cables i will appreciate since its original monitor is faulty RE: ATL - HDI 3000
Updated Jun 24 2020
2 RepliesBuenas tardes, tengo un problema con ecografo ATL hdi 3000 que no enciende , solamente prenden los leds 6 y 7 de la parte trasera, alguien me podra ayudar??
Muchas gracias RE: ATL - HDI 3000
Updated Jun 9 2020
4 RepliesHola, tengo un problema en ecografo ATL hdi 3000 que no se inicia, al encenderlo queda en stand by con los led 6 y 7 encendidos y al poner en on de la tecla lateral no hace nada, agradezco si me pueden orientar con esta falla.
Saludos
Walter RE: ATL - HDI 3000
Updated Sep 20 2019
11 Replies
Hi, After Boot screen apears normal, but an error mensaje show up about a Scan head problem, change to another Scan head and so on.. always the same error-.. Scan head initialization error ... i got the service manual. but i couldnt to find the board that is failing. help will be apreciate.
EDITED: thanks for all tips.
Finnally the solution to this problem was: change the AIM board located behind the Scanhead select
RE: ATL - HDI 3000Updated Aug 2 2018
3 Replies
my machine's screen not working what should i do ?? and if i want to connect an external monitor what is the connectors?? please contact me at my email : kmousa213@gmail.com .
RE: ATL - HDI 3000Updated Jul 13 2018
4 RepliesAI need help getting out of or exiting the scanhead selection screen view so i can image my patient. scanhead is attached and selected. how do i exit this screen? Trackball only rolls anterior & posterior also, will not go right or left RE: ATL - HDI 3000