I have GE Infant Panda Warmer that would not give an audible alarm for the heater or for the scale, but would alarm when the power switch in the back was flipped off. The warmer is now giving a system error 06 and give a red screen. Anyone have an idea of what is going on with the warmer. I appreciate any advice.
RE: GE HealthCare - Panda iRESHi all,
I have the panda resus hooked up to a liner and canister. When I turn the suction on, I'm expecting a vacuum of 50 mmHg but only measure ~20 mmHg. I also try increasing the expected vacuum but it still only measures ~20 mmHg. The vacuum does increase when the liner is manually pushed down when the suction is on, but manual intervention shouldn't be necessary.
The resus passed all checkout tests, I tried calibrating it anyways, and it passed all checkout tests again.
I checked for cracks or leaks in the liner and canister and didn’t notice any. Tried replacing the liner and canister with another set and that didn't resolve my problem. I also tried both the original set and the replacement set on another vacuum and they both worked as expected without manual intervention.
Any recommendations for what to try next? Tests, replacement parts, or something else?
RE: GE HealthCare - Panda iRESHello all,
The screen on my Panda will not light up.
I can see the screen when I shine a flashlight on it.
Is it just the backlight or is there something else involved?
RE: GE HealthCare - Panda iRESEquipment: GE HealthCare - Panda iRES
RE: GE HealthCare - Panda iRESI have a panda that the procedure lights intermittently turn off.
Equipment: GE Healthcare - Panda iRES
RE: GE HealthCare - Panda iRESI have a Panda warmer that the bed will not go up or down. I have tried the pedals on both sides, still no response. However, I when I press the pedals, I do hear a relay clicking. Please help.
Equipment: GE Healthcare - Panda iRES
RE: GE HealthCare - Panda iRESEquipment: GE Healthcare - Panda iRES
RE: GE HealthCare - Panda iRES