Shimadzu - MobileArt Plus MUX-100H
Shimadzu MobileArt Plus MUX-100H Forums
If you need help with a Mux 100 or Evo let me know. Worked on them for 8 years and have the info you may need. Just send me a message.
RE: Shimadzu - MobileArt Plus MUX-100HGet F14 error shooting above 80kv. 2 point cal works OK. Try 90kv get error. Tried each point cal from 40kv uo 80kv OK. Try 100kv get F14? Any thoughts?
Thanks
Sal Aidone, deccaid services
RE: Shimadzu - MobileArt Plus MUX-100HNeed a service manual for a Shimadzu MUX 100-H. We are getting an F14 error code. Any assistance would be greatly appreiciated
RE: Shimadzu - MobileArt Plus MUX-100H
When you turn the keyswitch on, the contactor is energized for ~1s than dropping out. If connected to AC, the machine is making constant high pitched sound. I do not have the service manual. Could somebody send me the manual?
My email is kkris1@yahoo.com
Many thanks!
Kris
RE: Shimadzu - MobileArt Plus MUX-100H
i can help please....i need service manual Shimadzu MobileArt MUX 100H
RE: Shimadzu - MobileArt Plus MUX-100Hi just repalced the batteries on this unit and now am getting this error and it is not in the manual i have any one know what it is?
RE: Shimadzu - MobileArt Plus MUX-100HI have an MUX-100H that wont charge. i have power to the T1 transformer but seems like nothing comes out of charger board. anyone have a transformer go bad on one of these?
RE: Shimadzu - MobileArt Plus MUX-100H