0 RepliesHello.
I have Bausch & Lomb - MILLENNIUM which does not reach the maximum of its vacuum value from venturi module.
The maximum value remains between 100 120 mmhg, even if I set the maximum slider to 500mmhg.
And sometimes the cassette does not want to eject, with a residual vacuum of 10mmhg, in this case I have to remove the source of compressed air so that the cassette is released.
No error code is present.
RE: Bausch & Lomb - MILLENNIUM
Updated Apr 26 2022
1 Reply The power supply module is in working.
There is problem in venturi module, the card of venturi module is not detected RE: Bausch & Lomb - MILLENNIUM
Added Oct 21 2021
0 RepliesDear sir!
Which connector should we use to connect a standard vitrectomy needle to the right port of the Enhancer MVE? Conventional plastic tips to the air port of Millenium System do not connect.
Thank you in advance, Boris. RE: Bausch & Lomb - MILLENNIUM
Updated Jul 5 2021
Updated May 29 2021
3 RepliesDear sir
Would you please help me to solve the device problem error 20 appear in the device screen and kindly provide me with the device service manual if it is available
Thanks & Best regards RE: Bausch & Lomb - MILLENNIUM
Added Feb 11 2020
0 RepliesDear Sir,
Need Service Manual of Millennium microsurgical storz cx6000(Bausch & Lomb)
Error(124):Surgical Mode Disabled due to missing/Uncalibrated Aspiration Mode.
Error(79): Module has failed to install.
Thanks for your corporation & time.
Regards
Muhammad Zohaib
angeloflove254@gmail.com RE: Bausch & Lomb - MILLENNIUM
Updated Mar 28 2018
Added Dec 20 2016
Updated Dec 29 2016
3 Replies
After damage the HDD in main system module I'm change HDD to new-one and reinstallation Windows 95 & Millennium v.2003-1 soft. But it's impossible to recover the "Doctor Settings" for Posterior zone (Total & Subtotal Viteroctomy). May be somebody can help me? I shall be glad to any help. I am in Uzbekistan and eqipmentis unique here.
RE: Bausch & Lomb - MILLENNIUMAdded Jan 22 2016
0 Replies
The averedge ultrasound power that appears after the phacoemulsification is in what unit?joules?