Siemens - Sireskop SD
Siemens Sireskop SD Forums
Need Service manual instructions for tube change on this product.
Can anyone tell me how could I enter to siemens sd 30 Tympanometry calibraton? Thanks in advanced.
RE: Siemens - Sireskop SDSome how the Barium Single shot protocol has been deleted from my system. How do add this back as an option?
RE: Siemens - Sireskop SDSome how the Barium Single shot protocol has been deleted from my system. How do add this back as an option?
RE: Siemens - Sireskop SDWhen angulating the overhead tube support to make xray on the table or wall bucky, the angulation displayed on the D23 board on the control console is incorrect. When the overhead is angulated to 90 degrees for the wall bucky, the display reads 132 degrees. When the overhead is set to 0 degrees for the table bucky, the display reads 42 degrees. This prevents auto collimation. All shots on the table and wall buckys must be manually collimated. The system will only auto collimate for the table when the actual angulation (not what is displayed) is -42 degrees. It will only auto collimate for the wall when the actual angulation of the overhead is 48 degrees. Any suggestion about how to approach this. Already tried replacing D23 board. No change.
RE: Siemens - Sireskop SD