RE: GE HealthCare - Definium AMX 700Hi all,
my device came up with an error this morning,
Unattached Inode 424481
Unexpected Inconsistency: Run FSCK manually.
It asks me to input my password for maintenance, but it does not accept my password. The only option I have is to type CTRL-D which shuts down the system.
Anyone here have any experience with this?
Equipment: GE HealthCare Definium AMX 700
RE: GE HealthCare - Definium AMX 700I have a problem with Display Controller PN 2409241, it´s blocking the power supply LVLE, when the connector from Display Controller is disconnected, the power supply LVLE show 5,+-15 and 24 volts, i´m looking for the E824-E3060VF(modular DC-DC converter) but i can´t get the datasheet(i think it´s the problem, maybe it has a short circuit)
Equipment: GE HealthCare - Definium AMX 700
RE: GE HealthCare - Definium AMX 700I'm getting an error message on my machine: "Table Image Detector communication link not established. Auto-retry in progress. Reset system if problem persists."
I've been informed that I'm getting this error because the IP address of the detector has been changed.
The machine has a wireless connection which I setup, and it was after this that I started getting this error.
I just want to confirm that the image detector requires an IP address. If it does how do I set it up.
Thank you.
Equipment: GE AMX 700
RE: GE HealthCare - Definium AMX 700The machine takes 20 minutes to reboot. Unable to add patient each time i reboot according to the message log it say machine "System did no reboot properly please reboot system" message. Please help
Equipment: GE Definium AMX 700
RE: GE HealthCare - Definium AMX 700