I am trying to reconfigure my Gemini Laserscope from 240v to 208v. I have flipped the switch on the back to 208 and have moved the labeled connection on the back right as well, but still get error 86. What other reconfig steps am I missing?
RE: IRIDEX - Laserscope GeminiOur machine is stating "PROBLEM 10". How do you fix problem 10?? It states possible causes as low coolant level, obstruction from broken glass, pump thermal shutoff/ failure and pressure/flow switch failure. HELP please :)
RE: IRIDEX - Laserscope GeminiReplaced the cooling system and would not power on after installation. Replaced the 24v power supply and now turns on with the error: Problem 22.
RE: IRIDEX - Laserscope GeminiMachine was serviced for PM in March 2014, and has never had a problem before that. I have owned it for 10 years. Called Cutera, and they think its the Q switch. Would like a quote on fixing this. Office is located in Plano, Texas. Thank you so much!