i have a problem with rotation gantry,a big noise in rotation gantry at the end of exam
RE: GE HealthCare - ProSpeedCT-Scan prospeed FII I have image artifact with out any Error >> the artifact is many ring
RE: GE HealthCare - ProSpeedI have massage error on CT Scan prospeed FII .
the Error : OGP did not receive any replay from JEDI in response to Sysconfig CMD nor JEDI Capaility requestCMD.
and Error LED on OGP BD is light and CONF red LED on KV CTRL V2 on JEDI model is light
RE: GE HealthCare - ProSpeedHi
I have problem on CT-Scan prospeed Fll.that problem at side tube X-ray part .
it has 2 bubble inside it
and it has noise on image and the machine don't working calibration
regards
RE: GE HealthCare - ProSpeedthis is ct scan prospeed Fll
it dose warrmup normally but can't do callibration.
I install SW agine but same problem.
I try do Q cal but it apear scan canseled and the error maseege apeare:
Q cal in remaining:2
q cal check:status=1'error channel=39,115,42,0 this is changed per try
error:189022
RE: GE HealthCare - ProSpeed
During Daily calibration the scans stop about halfway through and give the a calculation error.
Air calibration is ok
but during phontam calibration stop scan & Q cal stop scan
from error log
Qcal check : status= 1, error channel=0,68, ,39
RE: GE HealthCare - ProSpeed
my machine is CT Scan prospeed FII . i want software for it.
V/R-2.25(for v/R2.24)
part NO. 5197828
CTT 4.5.1
part No. 5144980
RE: GE HealthCare - ProSpeedI need the brushes for the slip ring where can I get it in Colombia or Panama???