GE HealthCare - Proteus XR/a
GE HealthCare Proteus XR/a Forums
Add Analog Radiography to My Bench23 Results
Updated Jan 24 2023
1 ReplyIs there a schematic available.
Equipment: GE Healthcare - Proteus XR/a
RE: GE HealthCare - Proteus XR/aAdded Aug 24 2022
0 RepliesThe OTS locks will not disengage. I have replaced power supplies, all boards related to the locks, OTS console. Anybody have any ideas? Thanks.
Equipment: GE Healthcare - Proteus XR/a
RE: GE HealthCare - Proteus XR/aAdded Mar 23 2022
0 RepliesHello.
My Proteus has a long delay between prep and exposure. This has been going on for several days now. No error is displayed and everything else seems normal.
Does anyone know what could cause this?
Thanks RE: GE HealthCare - Proteus XR/a
Added Sep 27 2019
0 RepliesGood morning,
We are interested in the DC Power Board, for the afore-mentioned equipment,
The part number is: PCB: 221-2875, ASSY: 2259298-44, REV.: 0, GEHL Medical Systems.
Also we are looking for the wall bucky buttons.
Regards,
Leonel Yeanpierre Rosales RE: GE HealthCare - Proteus XR/a
Updated Aug 15 2019
2 RepliesWhen I turn on the console it will boot up to the main screen, but I can't pick techniques, or select from the APR. All the squares that should have numbers or options inside of them are missing, all I see are empty boxes and no errors. I removed the cover of the generator and noticed that the IF board had no LED lights on, I only see 2 led lights on the COLL IF PCB and it is not the TX or RX LED. I measured voltages on the test-points and Im getting +12vdc , -12vdc, +5vdc and +9vdc. Could my collimator be the cause of the problem, or should I replace the Collimator IF board and PCB assembly in the generator? I noticed a day prior, seeing the main display on the collimator blinking on and off rapidly , but now is off. Should I focus on the collator instead? Thank you!!, Keith RE: GE HealthCare - Proteus XR/a
Added Oct 24 2018
0 RepliesWhen attempting to restore an APR backup, the console gives a "cer_ap.txt" and "zyg_arc.txt" error. These same errors can be reproduced by clicking on the associated buttons on the home screen.
I would prefer not to replace the console if this is an easy fix. Is this something that anyone else has run into in the past?
Thanks in advance for your input. RE: GE HealthCare - Proteus XR/a
Updated Aug 31 2018
Updated Jul 13 2018
1 ReplyTechnique and Parameters do not display at startup / remote doesn't work. I have disconnected all cables and reconnected. It will power on, but not display any numbers. RE: GE HealthCare - Proteus XR/a
Updated Jun 5 2018
3 RepliesHello, My Proteus XR/a System Console touch screen start to crack and less responsive. I wonder if anyone have experience of changing the touch panel before? What is the touch panel part number? and where i can purchase it?
Thanks in advance. RE: GE HealthCare - Proteus XR/a
Updated Jan 9 2017
3 Replies
When turning on my Proteus XR/a I get a message "Please Power Up Again" about 30 seconds after powering on. This unit does have a DR Upgrade. This is intermittent and sometimes you just need to reboot it a few times. Any ideas of what to look at? I am thinking the KV board. There are not any error codes logged.
Thanks,
Tim Deml
Rapid City Regional Hospital
RE: GE HealthCare - Proteus XR/a