GE HealthCare - Proteus
GE HealthCare Proteus Forums
GE Proteus - The light will not turn on because the switch is bad. When we hit the button it is only intermittently working. Can we just replace the switch or do we have to replace the collimator? If just the switch, do you have know a part number?
RE: GE HealthCare - ProteusI have an older GE Proteus Room where the SID to the wall bucky at 40" reads correctly, but at 72" it reads 63". It also reads fine tot he table. Someone told me that this is a simple fix that does not require logging into the system with Asian.exe. Would appreciate getting that information on how to do this. The person who said this did not remember how to do it, but said it was simple and done at the console. Thank you.
RE: GE HealthCare - ProteusI am new to this site and they had the old siemens collimator replaced with the Eclipse and we are having issue that the lamp will not come on after an exposure. this is something that is not happening all the time, but when it does they shut the unit down and reboot it and that clears the issue. I have replaced the OTS controller on the unit and looking at doing the I/F and KV controller in the Jedi.
Does anyone know if I am on the right track?
thanks
RE: GE HealthCare - Proteus
alternate source for Eclipse collimator lamp p/n 5494683-1 GE list price $1974. This collimator is made in Beijing, China.
RE: GE HealthCare - Proteus