Valleylab - Force 2
Valleylab Force 2 Forums
Updated Apr 24 2023
4 Replies
This Force 2 (Dash number 20CEM) will not output any monopolar power higher than 100 Watts. The power settings can be adjusted to the maximum values for each mode on the display (300 - pure cut, 250 - Blend 1, 200 - Blend 2, 150 - Blend 3, 120 - Coag). But when the footpedal or handswitch is activated, the unit will reset the power setting to 100 W, then output power at that setting. The Bipolar mode does not seem to be affected by this. This seems like a setting that needs to be adjusted. Any input is appreciated.
RE: Valleylab - Force 2Updated Dec 31 2022
2 RepliesHello Guys, I have a Valleylab force2 which works well but when I did the high frequency current leakage test, it did not pass the coagulation tests. It gives me values ??above the limit 150µA
If something similar has happened to you, it would help me to know what solution you gave it, thanks.
Equipment: Valleylab - Force 2
RE: Valleylab - Force 2Updated Apr 8 2022
1 ReplyI have an issue with the REM alarm. When I active the Coag mode, it starts to alarm like if the REM circuit was disconnected or something, and keeps like that until I switch off the equipments.
I read that it can be any of the IC but I tested them and its all OK.
If you have any other clue please let me know.
Cheers. RE: Valleylab - Force 2
Updated Dec 21 2021
Updated Dec 14 2021
2 Repliestengo un problema con el equipo: al momento de iniciar corte y en coagulcion se mantiene el sonido en intermitente al presionar corte y coagulacion si alguien le paso este problema porfavor me ayudaria mucho con su experiencia
RE: Valleylab - Force 2
Updated Sep 16 2021
1 ReplyHow is it going guys! I've a force 2 here and he's not working properly. He works just fine with hand switchs, however in the footpedal isn't, its activating the equipament, i can hear the sound, relay however nothing in the hand, no cut or cog RE: Valleylab - Force 2
Updated Aug 22 2021
5 RepliesI have a force 2 that works fine below 25, but if the power is set above 25 the unit will reboots. This happens in both coag and cut mode. Any suggestions?
The generator reboots if the power is above 26. RE: Valleylab - Force 2
Added Aug 22 2021
0 RepliesHello Guys, I have a Valleylab force2 which works well up to 26 W both cut/coag.
Any power above 27W will trip the generator with REM indicator and alarm constantly on.
I need your help . Thanks in advance. RE: Valleylab - Force 2
Updated Aug 7 2021
2 RepliesHello Guys,
I have a Valleylab force2 which works well with the Valleylab Non-Rem Patient Return Electrode (REF: E7506), but deosn't work with the Valleylab Rem Patient Return Electrode (REF: E7507).
Please I need help to make it work with the Valleylab Rem Patient Return Electrode (REF: E7507).
Thanks in advance. RE: Valleylab - Force 2
Updated Jun 16 2021