Im noticing baseline noise on my 3 lead ecg when the monitor is plugged into the outlet fed by the battery backup system. unplugged its fine, plugged into the anesthesia machine it sits on which is plugged into that UPS, its fine. I'm think ppwer supply, any other ideas?

RE: GE HealthCare - Cardiocap/5
Updated  Aug 10 2018
4 Replies

Hi anybody have the DRI Specification document?

RE: GE HealthCare - Cardiocap/5
Updated  Jun 8 2018
2 Replies

I have a Cardiocap/5 MON CONFIGERATION B FBWG02335 SN. The monitor reads 'SENSOR INOP'. The D-Fend and the battery have been changed. No Change. The Service Manual says "check timeout and bad checksum valves at the service menu" Not sure how to do that. Is this something I can do. When I call a the GE National Service Centre - they said "gas module needs changing" at a cost of 3500.00. Regards GL

RE: GE HealthCare - Cardiocap/5
Updated  May 2 2018
1 Reply

I have a cardiocap5 that I just replaced the NIBP cuff, hose and fittings. It will cycle up to 250 but then stops and comes back air leak. We have tried reconnecting everything but cannot clear it. Any suggestions?

RE: GE HealthCare - Cardiocap/5
Updated  Jan 15 2018
1 Reply

I need help from a biomed familiar with the Datex Ohmeda monitor in the Washington DC area.

RE: GE HealthCare - Cardiocap/5
Added  Oct 27 2017
0 Replies

There an error message "check sample gas out" I note that is no suction at the sample gas inlet. Any suggestions??

RE: GE HealthCare - Cardiocap/5
Updated  Aug 22 2017
2 Replies

Change CO2 from percentage to numeric value ?

RE: GE HealthCare - Cardiocap/5
Updated  Jun 21 2017
4 Replies

I just replaced the nvram on the cardiocap 5. Now, the monitor is not displaying data from the NESTPR. The menus are on the screen, but no data. I can't start a bp. The spo2 probe lights up, but does not display data. Any thoughts?

RE: GE HealthCare - Cardiocap/5
Updated  Jun 21 2017
1 Reply

I just replaced the nvram on the cardiocap 5. Now, the monitor is not displaying data from the NESTPR. The menus are on the screen, but no data. I can't start a bp. The spo2 probe lights up, but does not display data. Any thoughts?