I have adjusted the scale factor to it's highest rate and am looking for the next best possible reason the unit is under-delivering. Is it a finger pump or controller board issue?
RE: B. Braun - Vista BasicI am working on a BBraun Vista Basic and it is under-infusing. I have adjusted the scale factor as low as possible (400) and am looking for additional information. Could the problem be in the controller board, finger pump or both/niether?
RE: B. Braun - Vista BasicDoes anyone have a Braun Vista service key that we could purchase?
RE: B. Braun - Vista BasicFellow techs,
I am reaching out again for help with this wierd problem.
We have no service info on this thing, and BBraun isn't very nice if you haven't been trained by them.
So, I have this infusion pump that we can input the time and vtbi, but as soon as you hit the start button, it alarms and displays the following error...... Check drop sensor.
There is no drop sensor, so I'm very confused and in nned of some assistance.
I also need to know how to adjust the contrast on the display.
Any help is appreciated.
FOLLOW UP>>>>>>>>>>>>
Does anyone know where or how I can get a service key?
Is it an actual 'key' or is it software programming?
Thanks again.