Sysmex - XS 800i
Sysmex XS 800i Forums
I have
- Sysmex XS 800i
- Sysmex xn 1000
- Sysmex ca1500
and I need web service to Integrate with output data of these machines
RE: Sysmex - XS 800iPeriodically, during analysis of the same sample of a normal blood, there are bad leukocytes distributions along the vertical axis (SFL) of the DIFF scattergram occurred. When the mistake occurred, the distribution along this SFL axis is too short, and the flag «WBC Abn Scg» appears. In these cases (approx. 2 times in the series of 15-20 normal attempts), there is no leucocytes’ formula, or this formula is doubtful.
Also, when I measuring E-check control blood, the DIFF-Y value sometimes drops down lower than normal range.
What can be the source of this problem?
1. The laser block? (The laser oscillation time is 15843:36. The laser’s current 1,5/42,8 mA)
2. The FFS reagent dosing unit? (STROMATOLYSER-4DSDP count = 43136)
3. The power block?
RE: Sysmex - XS 800iHi.I have one XS-1000i but i have some HDD problems, so i need to format the disk, but i do not have the IPU software, so if you could help me with that i will appreciate all your help.
Kind regards
Alexander.
RE: Sysmex - XS 800iI wanted to connect my pc to network which is already connected with sysmex so can u please explain me how do i connect my pc to network with sysmex and also wanted to know how do i obtain sysmex xs 800i ip i forgot my sysmex ip while troubleshooting
RE: Sysmex - XS 800i1) PC whas reinstalled after hdd crash by using XS Recowery DVD
2) I werify that IP adress of PC is 192.168.28.150 and mask 255.255.255.0
3) software IPU 00-19 whas installed, with analyser serial number 15106 (correct number)
4) there is no communication between PC and analyser, Red LED on front analyser`s panel is on, but thats all.
5) ethernet link is UP, but no incoming packets are recieved. i use blue ethernet cable(it is "cross")
6) i can ping analyser`s adress by entering command ping 192.168.28.151, in this case i have recieved packets.
7) also i have trying install SUIT (in suit manager no connection also)
RE: Sysmex - XS 800i
How reliable nRBC USING SYSMEX XS 800i 5 PART DIFFRENTIAL ANALYZER?
Do one require special reagents or is it a part of regular CBC/DC count.
I read one requires XN model & special low ph reagent to do so/ is that correct
Haren
RE: Sysmex - XS 800iCurrently our instrument has patient surname showing their first name, and first name showing their first name on printout.
How can we reverse this setup,???
B
RE: Sysmex - XS 800ii need to send data from analyzer
to host computer but i do not know how to do this
please help me
RE: Sysmex - XS 800i