Siemens - Acuson SC2000
Siemens Acuson SC2000 Forums
Equipment: Service manual
RE: Siemens - Acuson SC2000Hello we have several Siemens sc2000 ultrasounds that we would like to install an extra protocol license on
some need a stress license and some need a vascular license
Is there any way to do this without having a contract with Siemens ?
Equipment: Siemens - sc2000
RE: Siemens - Acuson SC2000Hello. We have a Siemens Acuson S2000 device (2011) and a modular for it ABVS (2016).
Connection, detection and installation of software proceeded as usual. The system module sees, the sensor also detects the fabric and there is an image strip.
But when you try to press the button to start the examination, the sensor in the box only slightly twitches left and right, freezes and the program displays the error: “A System error occurred. Imaging has been disabled...”
And it must go through a full cycle from left to right.
It was tested before purchase and everything worked.
When the module is turned on, the system should also test itself by running the sensor along the entire line, but it also twitches slightly and freezes. No physical interference was detected - the sensor moves easily in both directions manually. We also checked the motor separately, connecting it to a laptop - it is working.
We need a specialist to help figure it out.
There is a suspicion that the software in the system (2011) is outdated and is not compatible with the ABVS (2016)
Please tell me what we can do?
RE: Siemens - Acuson SC2000Equipment: Siemens - Acuson SC2000
RE: Siemens - Acuson SC2000