Philips - M3046A M3
Philips M3046A M3 Forums
Updated May 12 2020
7 RepliesI am testing two Philips - Agilent M3046A patient monitors with M3000A servers attached with the Intellivue XDS suite and am running into a connectivity issue. Right now, XDS is installed on a Windows 10 PC (firewall off, 172.31.0.0/21 network segment), testing against version L.00 and N.01 to see if I can get the XDS software to connect with the monitors. Looking at the documentation, there should not be an issue but when I connect the monitors, I get "Monitor identity check failed: TIMEOUT [monitor name]". I can see the patient monitor in the Home Monitor list (trying one monitor at a time) and I can see the connection attempt to initiate but the connection drops. I have been looking for a compatibility matrix and additional settings I could change for the XDS platform (certificates, timeout settings, encryption levels, etc.) but I have not been able to locate any details. Any information you can share would be greatly appreciated! RE: Philips - M3046A M3
Updated Dec 9 2014
5 Replies
When the MMS Module is connected to the monitor it flashed pacer not captioning on the screen. Is this a main board issue or an ECG board issue?
RE: Philips - M3046A M3