CareFusion Alaris 8100 Forums

Updated  Jan 18 2022
2 Replies

Any idea what can cause this? or the FIX

RE: CareFusion - Alaris 8100
Updated  Jan 14 2022
2 Replies

Any idea? how to fix

RE: CareFusion - Alaris 8100
0 Replies

Hello, I am hoping someone can help me with my confusion regarding software versions for the Alaris 8100 and 8015 units. I have been getting "CHANNEL ERROR" on about 20 different 8100 pumps. I know this is a software compatibility issue. However, I am not sure which versions are compatible together (8100 & 8015). There are so many different versions on these pumps and it seems to be random. Here is an example: -8100 version 8.5.29.0 is compatible with 8015 version 9.12.40.4, but NOT with 8015 version 9.33.1.2 -8100 version 9.33.0.50 is NOT compatible with 8015 version 9.12.40.4 -8100 version 8.5.29.0 is compatible with 8015 version 9.5.32.2, but NOT with 8015 version 9.33.1.2 I need help figuring out which units to flash new software on and which version I need to be flashing. Thank you!!

RE: CareFusion - Alaris 8100
5 Replies

What repair needs to be done in order to fix a severely failing volume accuracy test? I replaced the Logic Board and Motor Controller Board. The 8100 is still only infusing 9.77g out of the programmed 12.0g. Should I replace the Motor? Pressure Sensors? Not much left after those...

RE: CareFusion - Alaris 8100
Updated  Sep 20 2021
2 Replies

Hello. I have a CareFusion - Alaris 8100 double-beeps when keys pressed at keypad test and doesn't register the keys being pressed on the test software. Just replaced Main PCB and updated serial number. It's only these new main pcb's that this happens. Any suggestions? Also, the 8100 doesn't power ON when FLASHING PUMP. The Flash Pump option times out because the 8100 won't power back ON along with the PC Unit...

RE: CareFusion - Alaris 8100
Updated  May 5 2021
2 Replies

What can cause this error code?

RE: CareFusion - Alaris 8100
Updated  Apr 29 2021
2 Replies

I have an alaris pump module (8100) displaying a solid "CH ERROR" and four dashes in the rate box. It doesn't scroll or flash. It also won't connect to the maintenance software. IUI connectors look good and it does this no matter which side it is connected to. Would this be a problem with the logic board?

RE: CareFusion - Alaris 8100
Updated  Feb 24 2021
2 Replies

Hello, I was wondering if anyone here within the community has ever had any complaints from clinicians about door latches on the Alaris 8100 Infusion Pumps being difficult to open. With and without the IV set in. Is it normal for the pumps to have a little resistance to them as you try to open or close? If you open the doors slow it is difficult to open. If you are quick and kind of flick the latch up the doors open fairly easy. In an older publication from BD it stated though that flicking the latch open is not a safe/recommended practice since it can cause fluid to potentially free flow. At least for the door latches that had a centered sear pin design. It also seems like all the replacement door latches no longer have the set screw adjustment on the Sear and now just have a plastic bump that was cast into the mold during production. So it does not look like it's possible to make any adjustments. Any tips or advice would be greatly appreciated. Thanks!

RE: CareFusion - Alaris 8100
Updated  Feb 3 2021
3 Replies

Alaris 8100 channel error 200.5040 replace channel with an operational unit as soon as possible. Any ideas what is causing this error?

RE: CareFusion - Alaris 8100
Updated  Sep 22 2020
2 Replies

During PM, you are asked to insert a dry set and close the door. After closing the door, i hit proceed and it keeps saying that the door is open

RE: CareFusion - Alaris 8100

Start a Thread

Equipment Name: CareFusion - Alaris 8100