I'm tring to locate a service manual for the GE OEC 9900 carm that has a parts list and hopefully a section for replacement procedures. Can anyone email me a copy? Thank you.
Keith
RE: GE HealthCare - OEC 9900 Elitedoes anyone know the video output resolution and refresh rate for this? ie 1280x1024 @60hz. and if it is 60hz, is it actual or like 59.97
RE: GE HealthCare - OEC 9900 EliteHi! Our equipment when we push the xray button, it shows a very scratchy and grey image and starts to shut down. This happens in long surguries when we have to give more that 140 images. Like if it gets hot and needs to shut down... I have pictures. contact me if you want to see them. ricardonunoamaro@sapo.pt
RE: GE HealthCare - OEC 9900 EliteI have a 9900 C-Arm, Of course this is intermittent!
When booting up system, sometimes it gets stuck and will not complete . The C-arm arrows get stuck at 5 arrows. The workstation just has the loading bar in the middle. Shut down try again and it works. Cable connections are good, Power cord is good. Another issue is Corruption Images Deleted error messages. Sometimes when saving patient images the Xray techs gets this message. The UPS is working correctly, boot-up sequence I verified. Do you need a curtain service laptop to login and check any logs ? Not sure what ArcNet is or Utility suite is ? This system is from 2011, any idea's ? Thanks
RE: GE HealthCare - OEC 9900 EliteDear all,
I am facing a problem with GE C-Arm model OCE 9900.
1- Some time system during startup doesent go to ready position.
2- when system gets ready for expose after every two or three exposes system is hanged.
Need Swift response.
Regards
jalal
RE: GE HealthCare - OEC 9900 EliteDear all,
I am facing a problem with GE C-Arm model OCE 9900.
1- Some time system during startup doesent go to ready position.
2- when system gets ready for expose after every two or three exposes system is hanged.
Need Swift response.
Regards
jalal
RE: GE HealthCare - OEC 9900 Elite
hello I have 9900 elitte during expose it become hang.no tach screen work.nothing work total system become hang.and the light above the touch become red.what is the problem in my unite.
RE: GE HealthCare - OEC 9900 EliteAre there any special requirements with regard to the room construction for using an OEC 9900 Elite Premium Digital Mobile C-arm? For example, lead shielding, etc.
RE: GE HealthCare - OEC 9900 EliteHello,
I have a GE OEC 9900 Elite that gives the error message...
"CORRUPTED IMAGES DELETED
Curruped images were detected at bootup and could not be repaired. These images have been deleted. This may be the result of a power loss or improper shutdown."
1. I was trained at OEC on this system a year ago.
2. I have performed the UPS checkout procedure and it passes.
3. I measured the battery voltage on each battery and all are above 13 volts DC. The batteries were on my work bench.
4. I timed the boot up and shut down of a known good 9900 and they both boot and shut down in the same amount of time give or take a second.
5. This problem happens every time the system is tuned off then turned on.
Thank you for your help...
RE: GE HealthCare - OEC 9900 EliteHello,
I have a GE OEC 9900 Elite that gives the error message...
"CORRUPTED IMAGES DELETED
Curruped images were detected at bootup and could not be repaired. These images have been deleted. This may be the result of a power loss or improper shutdown."
1. I was train at OEC on this system a year ago.
2. I have performed the UPS checkout procedure and it passes.
3. I measured the battery voltage on each battery and all are above 13 volts DC. The batteries were on my work bench.
4. I timed the boot up and shut down of a known good 9900 and they both boot and shut down in the same amount of time give or take a second.
Thank you for your help...
RE: GE HealthCare - OEC 9900 Elite