Philips BV Pulsera has no image when Flu., kv value goes up to max, sometimes error M303 "mA guarding failure" sometimes (most) no error.
I checked Image Intensifier and II power supply are good.
I changed the umbilical cord and the symptoms are still the same. Please advise.
Chao.
Equipment: Philips - BV Pulsera
RE: Philips - BV Pulserahello. i am getting error M 050 out of nowhere
i checked the cable it is working well with another Unit , tried to replace to swap the board and it seems to boot fine.
1-what shall i do shall i get the software reinstalled or rather replace the board ?
2- What do i need to download the software BV Scope ? how to find it ? (kimohaker@gmail.com)
3-Do i need ceratin key Software key or Hardware Key to get it done ?
4- Does it require any dongles for connection or only ethernet cable will be fine ?
Thanks