i would like to change the settings on one of the vents that are currently on pressure control mode. how do I get to the appropriate code
IK MD , Florida
RE: Datex Ohmeda - Aestiva/5 7100We have been comisioned to do maintennace on a 7100 machine, the first thing we found was a leak of 15 liters. We have replaced all the o´rings and preventive maintenance kits (as this machines have never been done so). The leak has droped to 500ml, but the manual indicates <200ml. Any sugestion on how coudl we look for were is the leak?. Machien passes all the tests after we have done a complete calibration. But when we setup the volumne like to 250ml, it measures 240ml or so, but the pressure increases to 30-31 cm H2O, and if we set up a volume of 500mL, the pressure goes up to 90 cm H2O. could this be a cause of the leak we still have? Any help will be greatly appreciated.
RE: Datex Ohmeda - Aestiva/5 7100Can you please send me the rate quote for this machine pls to my email eochoa@completedlogistics.com
RE: Datex Ohmeda - Aestiva/5 7100
some of our machinres have volume only mode. How does one change the ventilator to also use pressure mode? Thx, GGL