Busy signal appeared whenever it is used and Circuit error shown
Equipment: Draeger - JM-105
RE: Draeger - JM-105Hi guys,
I replace my batteries on a 3 year basis on these units as well as checking on the validity of the self-test measurements using the test disk under the charging base cover. I cannnot find any information on resetting the annoyance alarm when you turn on the unit stating "Time for periodic Calibration". I was told by Drager that calibration is only required if the user test of the three values falls outside of normal values. That and the battery being replaced on a schedule is the extent of the PM for these units. I do not have equipment to calibrate these units, so I rely on the operator self-check using the disk as the reference source.
Does anyone know how to reset the unit so the annoyance alert will not appear?