Physio-Control Lifepak 15 Forums

Updated  Feb 27 2024
1 Reply

I have a Lifepak 15 that loops around the self-test screen 3 times without fully turning on. When turned on the screen will show the self-test in progress screen seemingly shuts off then goes into the self-test screen again, this keeps happening 3 times and will just completely shut off after the third time. Batteries don't seem to be the issue as I have tested it both on batteries and AC. Any idea what could be the problem? PCB?

RE: Physio-Control - Lifepak 15
Updated  Dec 5 2023
3 Replies

I keep getting repeating error codes after self test on my machine was wondering if anyone has encountered this and was able to fix it. so far I have taken the machine apart and cleaned contacts on all the harnesses. I think its probably board level or software related. 

the codes are 

5006 

PC_ERROR_BAD_CHECKSUM (Program contents failed Checksum test.)


103B

UI_ERROR_MISSING_LANGUAGE (Configured language was not available.)


400B

SM_ERROR_CRC_FAILURE (Program contents failed CRC test; value = high 16 bits expected CRC, low 16 bits include computed CRC).)

RE: Physio-Control - Lifepak 15
4 Replies

Have to do Electrical safety test but do not have the Battery adaptors, what can I do?

RE: Physio-Control - Lifepak 15
Updated  Feb 6 2022
1 Reply

My STJ levels are not printing on the 12-lead and i can't find any info about it online.

RE: Physio-Control - Lifepak 15
Updated  Mar 3 2021
2 Replies

Any recalls announced ?

RE: Physio-Control - Lifepak 15
Updated  Apr 8 2019
3 Replies

Printer is printing enlarge print and majority of ECG is off the paper. Is there a setting to change or does it need to be sent to physio for repair?

RE: Physio-Control - Lifepak 15
2 Replies

What would cause the discharge levels to read high. I am getting high readings when testing on my analyzer. For example on 360J I get a reading of 373.9J. Thank you.

RE: Physio-Control - Lifepak 15
Updated  Jul 13 2017
2 Replies

What would cause the ECG Lead II printed signal amplitude to exceed the 36 mm to 44 mm peak-to-peak tolerance? The values I achieve on a REF 99577-001256 model LifePak 15 are 55-56 mm. Thank you.

RE: Physio-Control - Lifepak 15
0 Replies

What would cause the Shock # marker and Energy Delivered event marker to not be recorded on the Paddles printout?

RE: Physio-Control - Lifepak 15
Added  Jul 13 2017
0 Replies

I have a couple 2011 units that the displays read "Check SPO2" sensor no matter which SPO2 cord is attached. Does any magic non-repair fix exist for this error or will a new sensor be required? Thank you.

RE: Physio-Control - Lifepak 15