When monitoring patient, monitor is turned on and a first vitals reading is taken with no problem.
As second reading is then attempted after 10 minutes and the monitor blacks out and will not light up again.
When monitoring patient, monitor is turned on and a first vitals reading is taken with no problem.
As second reading is then attempted after 10 minutes and the monitor blacks out and will not light up again.
Check whether power is coming to the Display board. If 'NO', it is the connection cable from the power board to the display board or the power board itself.
Answer provided by F. Kareem, Clinical Engineer